Dear colleagues,
I found a problem in the ISO MME database and also the figures for the Y7 dummy. For the Femur there are two Load Cells given there:
one for the Lower Femur:
??FEMRLE00Y7FOZ?
??FEMRRI00Y7FOZ?

and one 6 axis for the upper Femur:
??FEMRLEUPY7FOX?
??FEMRLEUPY7FOY?
??FEMRLEUPY7FOZ?
??FEMRLEUPY7MOX?
??FEMRLEUPY7MOY?
??FEMRLEUPY7MOZ?
and the same for RI

But that seems to be wrong! There is just one location and load cell on each Femur. The Location seems to be more lower than upper.
Here you can have a one axis load cell or a 6 axis one.
Y7_LoadCells.pdf
(231.1 KiB) Downloaded 682 times
So we need to change the codings in the database and the figure to:
??FEMRLE00Y7FOX?
??FEMRLE00Y7FOY?
??FEMRLE00Y7FOZ?
??FEMRLE00Y7MOX?
??FEMRLE00Y7MOY?
??FEMRLE00Y7MOZ?
and the same for RI
The "UP" variant must be removed.